Wavy baseline on agilent 1100 VWD

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Column: C18 Varian
Flow 1ml/min
Gradient: First 5 minutes: 85% ACN and 15% water
next 25 minutes: 100% ACN

Everything was fine, UNTIL I switched mobile Phases from:
Water (MPA) and Acetonitrile (MPB)
To:
Water w/0.1% TFA (MPA) and Acetonitrile w/0.1% TFA (MPB)

I mixed them both well. I already replaced the lamp, that did not help.

Before I replace the pump seals, does anyone know what else I can try?

Should I flush the column with Methanol? Reverse flush the column to elute a late contaminant peak?

Thanks

TFA is notorious for baseline problems. See this thread, for example:
http://www.sepsci.com/chromforum/viewtopic.php?t=8447
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 27 users online :: 6 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Google [Bot], Google Adsense [Bot], John Guajardo, Semrush [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry