Uses a 30m x 0.32 x 0.5um Wax column, split injection (2.5 to 1), 0.5 uL inj vol., injection solvent is 100% ethanol (cannot change it due to extraction issues), column constant flow 3.0 mL/min of N2.
Analyte of interest elutes at 10.5 minutes, and validation performance was excelent. (RSD < 0.5%, Tailing 0.95)
Tried to transfer method to another lab that only has PE Clarus 500 GCs. With above parameters, peak elutes at 7.0 minutes and fronts badly (column overload). Eventually changed flow to 1.5 mL/min and achieved 10.3 min retention. Split ratio increased to 20 to 1 to minimize fronting (tailing 0.85).
Still having repeatability problems (RSD>5%).
Has anyone seen 6890 to Clarus 500 method transfer problems?
Any ideas as to the cause the drastic flow and split changes necesary to achieve similar chromatography?
(We've ruled out the typical hardware and column installation issues.)
The injector geometry is slightly different between these 2 instruments, so perhaps that explains the split difference, but for the life of me I have no idea why the flow needed to be cut in half!
Thanks,
Jon