Re: Viscotek TetraDetector--no light scattering peaks.

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
For GPC analysis we use a Viscotek Tetradetector array that has UV, RI, Viscometry, and light scattering detecors (right angle and low angle) with a THF mobile phase. I had a power outage and the instrument came back up but when I inject our polystyrene standards I see peaks in the RI, UV, and on the Viscometer signals but nothing on the LS signals. There appears to be some random noise but no peaks at all...

I can't see anything overtly wrong with the instrument...anyone have some ideas here??

Thanks for your help.
My guess is that the light scattering source laser is not outputting light for some reason (Laser on the TDA not turned on, or laser in need of replacement).

If you will contact our technical service department in Houston, we will be happy to help troubleshoot this for you.

You can reach the office at 281-445-5966, or visit the website at www.viscotek.com for information on technical services.

Regards,

Todd

_________________________________________
Todd Grey
Southeast Account Manager
Viscotek Corporation, A Malvern Company
704-792-1945 Office
800-648-6950 Toll Free
800-375-5966 x 203 Voicemail
wtgrey@viscotek.com
_________________________________________
_______________________________
Todd Grey
Southeast Account Manager
Viscotek Corporation, A Malvern Company
704-792-1945 Office
800-375-5966 x 203 Voicemail
wtgrey@viscotek.com
_______________________________
Right you are...
The Laser button was turned off. I must have had a brain spasm...

Thanks;
IC_Gal

Glad you were able to get it working!

By the way - the laser has a limited lifespan and will need replacement eventually. To make it last longer, you can keep it turned off except when running samples.

Thanks,

Todd
_______________________________
Todd Grey
Southeast Account Manager
Viscotek Corporation, A Malvern Company
704-792-1945 Office
800-375-5966 x 203 Voicemail
wtgrey@viscotek.com
_______________________________
4 posts Page 1 of 1

Who is online

In total there are 28 users online :: 6 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Google [Bot], Google Adsense [Bot], John Guajardo, Semrush [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry