It's also possible that the retention gap ( in front of the column ) needs replacing. Was the decline gradual or sudden?
If gradual, you may also be able to condition the column at temperature to reduce the activity that is destroying the triglycerides, but ensure the following first....
If sudden, it's really important to ensure that you have good quality, well maintained gas purifers on your carrier gas lines, and you protect the column from trace oxygen by having good cylinder changeover regimes - if you use cylingers of gases.
I strongly recommended that any laboratory performing routine QC have replacement columns, retention gaps, injector inserts. When ther is a problem, then the new column can be sustituted once simple problem solving suggests the root cause is the column.
Please keep having fun,
Bruce Hamilton