Danko, The analysis is of oligomers extracted from polybutylterephthalate, i know there are some acid compounds eluted at the start of the chromatogram.
That would be a plausible explanation of what you’re experiencing.
What the PH is i don't know. What would be used to adjust this?
My favorit choice (for a start) would be 0.05 M Sodium Phosphate, pH 2.3 – 2.5 (that is instead of the pure water in your mobile phase).
Is it safe to use a low ph with all columns?
Unless you’re using some kind of very exotic column/stationary phase, it is perfectly (100 %) save!
I’m assuming, the column you’re using is any kind of silica based C18 or C8 or something like that.
In my opinion the low pH buffering offers a number of advantages, especially in the case of acidic compounds separation on silica based reversed phase columns.
BTW, expect longer retention time for this/these peak/s.
Best Regards