Hi

i have been doing certain comparison studies…recently I analyzed a protein in porous (300Å) column and a non-porous column….
In non porous column under same gradient % per minute (around 1% ) the protein gave very wide peak..under lower % gradient the peak width went on increasing

Non porous column peak width 1.503 min (1% gradient ), porous column :- 0.647 min (1% gradient )..to achieve peak width of porous column I have to run under 3% gradient

Why is there such a drastic difference in behavior of non-porous and porous columns???

Thanks

The most likely cause of your observation is the fact that you overloaded the non-porous column more than the porous column. A non-porous column has a very small surface area compare to a 300 A column.
2 posts Page 1 of 1

Who is online

In total there are 22 users online :: 4 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot], John Guajardo, Semrush [Bot] and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry