I'm running a method which I am transfering from HPLC to UPLC and I have a wavy baseline during the end of the run (ACN increase) the thing is i don't get the same thing in HPLC, detector filtring didn't help and also not washing the system and the column.
what could be the reason for this and how can i get rid of it?

One explanation could be an insufficient mixing. With your HPLC system, this is probably better. However, it would be beneficial to know your gradient conditions to judge it better.

Yeah that's true !

We must know your final conditions of your gradient. I suppose you observe it when the gradient is the most elutive. With my own experience in UPLC, you could see a wavy baseline dur to a mixer issue. UPLC is equiped with a 50µL mixer. Try to use a 100µL mixer for example.
Secondly, sometimes you could see noisy peaks due to solvent impurities that you 've never seen in HPLC.
Try to use advanced grade of solvents like in LC/MS.

Have fun !
3 posts Page 1 of 1

Who is online

In total there are 23 users online :: 2 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: John Guajardo, Semrush [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry