danko » Fri Jun 06, 2008 11:30 am
Hi Adrien,
I’m missing some important information. I didn’t ask of course so it’s my fault.
Anyway, after the appearance of tailing peaks, is the column dead/useless or does it work fine again next time you start a sequence?
If it is the case, then you shouldn’t worry about precipitations and rinsing – especially if you are planning to use 100 % ACN.
It would only mean that you’ll need to prolong the re- equilibration step. I don’t know how long you equilibrate before the next injection, but considering the column size and the nature of the mobile phase, I would suggest at least 10 – 12 min re- equilibration step following the gradient.
If on the other hand the column is ruined after the 15th injection, then you might consider precipitations – not necessarily of the mobile phase components, but rather sample components in combination with the mobile phase. The quick test is, to mix mobile phase and sample in a test tube and observe the mixture for changes.
Finally, I still think your working conditions generate too much backpressure which with time might cause void volume at the head of the column, which will manifest in distorted peaks etc.
The solution in this case might be changing the MetOH to ACN or mixing MetOH with ACN.
It should be mentioned here that ACN has stronger elution properties than MetOH, so it wouldn’t be a simple substitution. Possible selectivity alteration should not be counted out either.
Best Regards