In my uv detector I see opitions signal polarity and data rate .how does it effects the chromatography
-
- Posts: 405
- Joined: Wed Dec 05, 2007 4:55 pm
Polarity refers to the output of the detector - do you want the peaks to go "up" or "down." This is a common option for refractive index, but not absorbance. Very few chromatographers use methods that generate negative peaks with an absorbance detector.
The data rate, or sampling rate, tells the detector how often to collect data (1 data point/second = 1 Hz). In general, you need to collect a minimum of 25 - 30 data points across your peak, to ensure that you get the best accuracy. So, determine the width (in seconds) of your most narrow peak, and set your data collection rate so that you make at least 25 measurements in that time. For example, if your peak is 15 seconds wide, you should collect data at about 2 Hz or faster.
The data rate, or sampling rate, tells the detector how often to collect data (1 data point/second = 1 Hz). In general, you need to collect a minimum of 25 - 30 data points across your peak, to ensure that you get the best accuracy. So, determine the width (in seconds) of your most narrow peak, and set your data collection rate so that you make at least 25 measurements in that time. For example, if your peak is 15 seconds wide, you should collect data at about 2 Hz or faster.
Merlin K. L. Bicking, Ph.D.
ACCTA, Inc.
ACCTA, Inc.
-
- Posts: 104
- Joined: Sat Apr 28, 2007 12:49 am
Thanks a lot
-
- Posts: 104
- Joined: Sat Apr 28, 2007 12:49 am
If I do not collect more than 25 data point for one peak..I mean i collect only 10 data point for my 15 sec wide peak How it will effect the peak shape?
-
- Posts: 405
- Joined: Wed Dec 05, 2007 4:55 pm
There are several problems that can happen when you do not collect enough data points. First, your accuracy will not be very good, because noise in one data point represents 10% of your peak data. The integration start and stop points will also affect the area and height. Now, these errors will not be 50%, but they could easily be 10% or more. So if it is an important analysis, I would recommend collecting more data.
Also, with only 10 data points, you can miss important information, like changes in peak shape, or the appearance of impurities or shoulder peaks.
Another consideration is how the software and hardware work together. On some systems, the data collection rate and detector time constant (response time) are linked together. A slower collection rate usually means a larger response time, which means that you could lose much peak information (including shape, width, height, etc.).
Also, with only 10 data points, you can miss important information, like changes in peak shape, or the appearance of impurities or shoulder peaks.
Another consideration is how the software and hardware work together. On some systems, the data collection rate and detector time constant (response time) are linked together. A slower collection rate usually means a larger response time, which means that you could lose much peak information (including shape, width, height, etc.).
Merlin K. L. Bicking, Ph.D.
ACCTA, Inc.
ACCTA, Inc.
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
Essentially, when you do not collect enough data points, your peak will be distorted, and you loose resolution to neighboring peaks. This is the main issue about an insufficient sampling rate.
6 posts
Page 1 of 1
Who is online
In total there are 27 users online :: 6 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Google [Bot], Google Adsense [Bot], John Guajardo, Semrush [Bot] and 21 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Google [Bot], Google Adsense [Bot], John Guajardo, Semrush [Bot] and 21 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science