3 posts Page 1 of 1
Hello,

I am running a GC of my kerosene samples in CS2. The kerosene concentration is 2% w/v in CS2. The peak of CS2 however, is very small almost half that of the highest peak of the hydrocarbon component of kerosene. I have not encountered this result in my previous runs. Is this a problem? CS2 which is present in higher amount and is supposed to yield an enormous peak.

My GC conditions are the following:
oven program: 30C for 4 min, 10C/min to 260C
Injector at 325C
Detector at 350C
The column is an Inert CAP of GL Sciences with an ID of 0.25 mm length of 30 m, df of 0.25
Im using Gas Chro 4000 of GL sciences

FID has very low response to CS2. The FID response is more sensitive to compounds containing C-H bonds. What you are observing is right and we use CS2 as solvent in many of our analysis for just that reason.. On the other hand, you cannot use CS2 as solvent when using ECD detector because of very high response which might temperorily create ECD response problem.

Cheers!
Suresh.

Thanks
3 posts Page 1 of 1

Who is online

In total there are 66 users online :: 0 registered, 0 hidden and 66 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 66 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry