Hi all,
Have you used these 2 columns (or similar) for alcohol analysis?  We are going to use n-propanol as internal standard and analyse ethanol content in food and wine samples.
We will prepare the standard solution by water (and also sample preparation).  If I inject 1ul of it, will water kill the column easily?  Some people say water will hydrolyse the PEG column and cause bleeding.  Is it true?
Any better suggetions are welcome.  Thanks a lot!
We actually use both DB-wax type and DB-624 capillaries for ethanol assays in hand sanitizer products.  The 624 capillary is same as specified in USP-611, which is the procedure we use, and yes, our QA still made us validate it anyway, as they said USP was for ethanol solutions, and our products have some other ingredients.  That USP 611 (and us) detail acetonitrile as internal standard (although we did also do with n-propyl alcohol, which worked fine under same USP-specified conditions.  This is a wide-bore capillary and utilizes a 0.5ul split injection.
We formerly used the wax-type column, also 0.53mm wide-bore capillary, and that assay utilizes n-propyl alcohol, the logical internal standard.  But my company agreed that it would be better to move to USP, which we did.  That also used 0.5ul split injection.  We used a 5ul syringe for both.