temperature and baseline

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Dear colleagues,

I have some issue with baseline. While temperature program is isothermal I observe increasing baseline. When the ramps starts baseline goes down and it is the same until run ends. Here you can see it on this picture. Have you any ideas what could I do to prevent baseline increasing at the begining of the run?

The temperature is 60°C until 4.5 minutes. Maximum temperature is 260°C. Blue line is from ECD and red one from FPD.

Image
Check for an air leak, it could be drawing in a little air at the low oven temperature then sealing once the temperature increases. You may have a cracked ferrule either at inlet to column or inlet to detector.
The past is there to guide us into the future, not to dwell in.
Is this an Agilent system? Initial baseline seems really high for a uECD, and even more so for ECD. Do you have a blank plug to check detector without the column?

Maybe oven ramp is chocking off column flow. Have you done any inlet maintenance, septa or liner?
3 posts Page 1 of 1

Who is online

In total there are 16 users online :: 0 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry