AICMM » Wed Apr 30, 2008 1:28 pm
bedlamVR,
First, use air as your un-retained peak. I am pretty sure there will be a small amount of air left in your pyrolyzer before you turn the heat or valve and this should be the first thing off, right? In fact, I would start with a low EM setting the first time I did this since I would be uncertain about the response. If at first I did not succeed, I would try again with a fresh sample and a higher initial EM. Or... simply inject into the septum on top of the cut into injector nut and measure the air peak retention time. Then you calculate the column flow. Second, the 5890 is back pressure regulated so if the head pressure gauge has not changed you should be getting the same flow (in theory.) Third, reduced sensitivity speaks to an increase in the split flow rather than a change in the column flow and that you should be able to measure off the split vent port. I am assuming in all of this that you are running split the whole time?
While I am at it I would trim the column (~20 cm) and replace the liner and septum. I would also question the condition of the transfer line. Again, an injection directly in the injection port should enlighten you on this, perhaps an alkanes mix C10-C14 or C10-C28, available from most supply houses. Plus, we are dividing the system in 1/2 which is always a good way to troubleshoot.
Best regards.