Could you provide more detail on how the peaks are changing?
Did this happen suddenly, or has it been happening slowly over days/weeks/months?
Suddenly.
Are the "bad" peaks tailing more than the good peak?
The "bad peaks" have a tailing factor of about 0.64. The good peaks have tailing at about 1.0 to 1.2. The bad peaks are fronting very badly, but the column is o.k.
Is this an isocratic or gradient separation?
Isocratic
Please specify your Agilent system (pump, detector, autosampler) and mobile phase components.
Agilent system A1100:
G1322A Degasser
G1311A Quaternary pump
G1313A Autosampler
G1316A Column Compartment
G1314A UV Detector
Mobile phase 0.05M KH2PO4 pH=3.0:Methanol (60:40). Flow rate 1.0ml/min, column temp 40 deg C. Leeway of +/- 2% for methanol, +/- 0.1 pH units for buffer, 0.1ml/min flow, =/- 5deg C column temp.
I've attached jpegs of the chromatograms:
The black trace is from the Agilent system.
This isn't the only setup that gives problems with resolution. Some of my colleagues have mentioned this may be due to the fact that the Alliance systems have a larger dead volume, so I will investigate that further, but in the meantime any input would be appreciated.