Patrik,
you can repeat some simple experiments that will show that your ZIC-HILIC is in your group
" Charged - strong electrostatic interactions
Plain silica phases, aminopropyl phases":

See if you can elute a positively charged entity (as much as micro mols) with a mobile phase consisting of 0.00001M strong aqu. base on your ZIC-HILIC.

Dear Hans,

ZIC®-HILIC Separation of Betaine from Chinese Lycium
http://www.sequant.com/sn/ufiles/SeQuan ... 00-30A.pdf
Mobile Phase (v/v): 75%, Methanol, 25%, Ammonium Acetate 10 mM, pH6
(total ionic strength 2.5 mM)

ZIC®-HILIC Separation of Choline and Inositol
http://www.sequant.com/sn/ufiles/SeQuan ... 00-40A.pdf
Mobile Phase (v/v): 80%, Acetonitrile, 20%, Ammonium Acetate 100 mM; pH 4.5
Overall ionic strength: 20 mM
The second example is at double the requested ionic strength, I know.

These are examples of positive entities eluted isocratically with reasonable retention factors (3-5) available at http://www.sequant.com/applications
Merck SeQuant AB
www.sequant.com

Hans,

I don't know what the problem is...

I got perfect peak shapes for morphine and its glucuronide on Atlantis HILIC silica with 10 mM ammoniumformate, pH 3. The separation was done with a gradient on a 5 cm column, and there was so much space between the two peaks that you could put several HPLC instruments between them.

Lessons:
1. You get good peak shapes for basic compounds with HILIC silica.
2. HILIC is a good technique for analyzing metabolites that are difficult to retain on a C18.

Again, I am not making any qualifying statements about silica in HILIC. I am just stating that I could not find a difference in silica interactions between ZIC-HILIC and Atlantis. If I remember correctly, I even stated above that under the same ammonium salt conditions for both columns you obliterate the SiO- interaction in both columns. Patrik´s and Uwe´examples nicely show this. I added the example of 0.00001 M base for the mobile phase under which one will not elute a positively charged ion from both columns.
My problem is how one can claim that ZIC-HILIC is unique in that it does not give SiO- interactions.
19 posts Page 2 of 2

Who is online

In total there are 23 users online :: 0 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 23 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry