Baseline problems -Acquity (Concave and Convex)

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

24 posts Page 2 of 2

Make your weak wash 1:1 methanol/water.

The problem could be the 3ul injection - why so small? try filling the 10ul loop completely and see what happens.

Are you using different autosamplers?
No Tswett

Hi Adrian,

The method has been validated with 10µL. We are in an overload posistion to gets sens for imps. If we injected 10µL we would lose Rs for peaks on the tail of the mainj peak - there is a very fine line for this method unfortunately.

I will get back with further results / conclusions when I have time next week. I think I need to follow up on the needle wash idea to be honest and then get back to this forum.

If the method has been validated at 10ul I think you should stick with it. if necessary dilute your sample. I am all for varying a method but even to me that is a 'bridge to far'!
No Tswett

Sorry my mistake, its should have read the method has not been validated for 10µL.

The injection volume of the method is 3µL and this has been fully validated to ensure Rs is assured when running the method.

A colleague of mine has run another method with AcN gradient - I'll let you all know the outcome in due course.

With regards to just the smile curvature of the 1st chromatogram (10 mm flow cell) it appears that the light guided flow cell is definitely at fault here. It is even explained in the Waters publication regarding use of their Median Baseline Filter.

I have done some further runs with the high sensitivty flow cell and the gradient profile is as expected - i.e. steadiliy rising rather than just convex. I'm not sure what has solved it, but a user before me was running AcN grads where before I was running just MeOH. Also I've changed the needle wash solvents to:

Weak needle wash - 50% water 25: AcN 25 % MeOH
Strong needle wash - 50% MeOH and 50% AcN

The sensitivity check of my method is now visible!

I will check the analytical flow cell tomorrow (and perhaps run some AcN grads down it) to see if I can get rid of the concave baseline affect with this cell as well. I'll keep everyone posted how I get on....

With regards to just the smile curvature of the 1st chromatogram (10 mm flow cell) it appears that the light guided flow cell is definitely at fault here. It is even explained in the Waters publication regarding use of their Median Baseline Filter.

I have done some further runs with the high sensitivty flow cell and the gradient profile is as expected - i.e. steadiliy rising rather than just convex. I'm not sure what has solved it, but a user before me was running AcN grads where before I was running just MeOH. Also I've changed the needle wash solvents to:

Weak needle wash - 50% water 25: AcN 25 % MeOH
Strong needle wash - 50% MeOH and 50% AcN

The sensitivity check of my method is now visible!

I will check the analytical flow cell tomorrow (and perhaps run some AcN grads down it) to see if I can get rid of the concave baseline affect with this cell as well. I'll keep everyone posted how I get on....

It was definitely a flow cell problem. A waters engineer came in and replaced the 10 mm analytical flow cell and "voila" case solved - expected gradient baseline rise from my method - something to watch out for methinks!

Hi Rob,

What was wrong with the flow cell that was replaced?
Also, what caused the wavy baseline?

Thanks for sharing.

Best Regards
Learn Innovate and Share

Dancho Dikov
24 posts Page 2 of 2

Who is online

In total there are 25 users online :: 0 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry