m/z drift

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

6 posts Page 1 of 1
I am using multiple reaction monitoring (MRM) to quantify compounds with Agilent LC/ion trap XCT Plus MS. I found that if the sample's concentration is very low and I set a couple of transitions (not single one), sometimes the obtained m/z in the acquired spectra changed. For example, I set transition m/z 309-->180, the resulting data got m/z 179. From the retention time, this peak is consistent with my desired compound. But why not showing m/z 180 as supposed? It this normal? Thanks.

Have you done a mass calibration of the detector?

yes, the mass detector was calibrated two weeks ago. Such drift only occurred for a few particular transitions.

Have you used another transition before this that has about the same parent or daughter ion? I am trying to imply somekind of cross-talk which is very common in triple quadrupoles but it might not occuring at all in the case of ion traps...

Hi, Kostas, take the above m/z 309-->180 as example, I also set m/z 309-->179, the retention time is same for both transitions. I can not recall the peak intensity for these two transitions, but think they seem to be close. For standards with higher concentration, I always got m/z 309-->180 which should be. That is why I am wondering if low concentration sample can result in this ("cross-talk" as you mentioned). I don't have experience to use triple Q.

Hi James,

I think I talked too fast, cross talk is inherent to quadrupoles (it happens when you follow transitions one after another where the daughter ion is the same and the fragmentation cell didn't have the time to get rid of the previous daughter ions) and not the ion traps. It is also independent the abundance of the ions...

At this time I do not see any good explanation why this is happening... if you have already tuned the isolation and fragmentation elements of your ion trap...
6 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry