anion exchange method for cu-edta chelate?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
hii.. does anyone have an anion exchange method for cu-edta chelate determination and quantitation?
i have a weak anion exchange column to use.. any methods for that? or anything else?
Explore!

i;d try this.. why H2SO4 by the way? only pH? or to avoid any other acid's use? why so?
why methanol for ion exhange? i would be using only anion exchnge column.. without an rp base.. so i need not use methanol right? thankss!!
Explore!

you need ions in the mobile phase to facilitate anion-exchange and elute EDTA from the column. Strong acid will help you obtaining better peak shape. You can use any acid or buffer but sulfuric acid and TFA will work the best. I would stay away from methanol for this particular application.
4 posts Page 1 of 1

Who is online

In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry