- Posts: 10
- Joined: Wed Jan 19, 2005 6:24 pm
since the late 90's. It has an auxiliary flow panel and a
nitrogen-phosphorus detector. I have been warned about
potential problems with the detector, but a gas flow problem
has stopped me before I could even turn on the detector
The gases were already plumbed as indicated in the manual,
so I have installed new traps and pressurized the gas lines.
The gases are carrier and detector makeup (He), as well
as H2 and Air. The auxiliary panel gauges indicate that the
gases are reaching this panel acceptably. Unfortunately, I cannot
seem to get any flow to the detector. Even after turning the
mass flow controller many times counterclockwise, and the
backpressure regulator many times clockwise, column head
pressure has remained at zero, and no flow has reached the
detector (according to my bubble meter). I took apart the
tubing directly upstream of the front flow panels, and there is
plenty of pressure.
So I wonder, is there a blockage in the front panel valves or gauges,
or have I in my ignorance of 5890A construction missed something
obvious, like a "turn all gas flows ON" switch?
thanks in advance for your comments!