but the side by side run on both columns turned out the peak eluted almost at exactly the same time(8.5 min.) from both columns, the general gradient is 90/10 water/ACN with 0.1%formic acid to 5/95 water/ACN in 15 min..
since the new column is almost 50% higher carbon load than luna, theoretically it should get more retention when all other conditions fixed. the vendor tech support can not explain this either, just suggest to run isocratically., saying that the gradient steepness may overcome the retention, that i doubted not true.
the compound is small drug discovery molecule(mw=450) with several secondary amine and amide structure, the peak shape is perfect on luna with acidic mobile phase.
i even tried another column claimed better for basic compounds from the same vendor, but it did not elute later than Luna C18(2) did.
so please suggest some highly retentive column for small molecule, i remember some vendors list charts of various columns comparison based on retention, polarity, and other criteria, but i can't find them now.
my purpose is to increase the organic content for signal sensitivity in LC/MS, since higher organic(less water) will help to increase ionization in MS.