non linearity

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
dear all,
we are doing analytical method development for dicycloverine hcl using lc method. when i inject 20 to 80 microlitre of my stock , i get good linear response but when i make a serial dilution the stock i cant get the lineariyu at all. the funny thing is, when i check with lowest concentration by injecting 10,20, 30 to 80 microlitre it shows linearity. i am wondering what went wrong with me. i check the dilution but it was correct. can anybody find this problem before.
with regard
Donal

It can be a number of factors including bad pipeting (instrument or human error). Have you calibrated your pipets? Can you describe the way that you are making the dilutions?
dear all,
we are doing analytical method development for dicycloverine hcl using lc method. when i inject 20 to 80 microlitre of my stock , i get good linear response but when i make a serial dilution the stock i cant get the lineariyu at all. the funny thing is, when i check with lowest concentration by injecting 10,20, 30 to 80 microlitre it shows linearity. i am wondering what went wrong with me. i check the dilution but it was correct. can anybody find this problem before.
Dear
I think Kostas is saying correct. Also so many factors such after preparing solution where your storing it or immediatly your keeping in autosampler it is nothing but you have to check stability of your drug in diluent also.

Also concern about the pipette.
All the best

On several occasions it has been pointed out that serial dilution should not be used to prepare standards. Any errors are multiplied as you increase the dilution. Dilutions should all be made by taking various volumes from the stock standard.

Serial dilution may be OK for homeopathy but not HPLC!
No Tswett

We would need more information about what you mean by "non-linearity." Is the response non-linear at high concentrations or low concentrations? Also, injection different volumes in LC can cause linearity problems as well. It is unusual, however, to see a better result with different injection volumes.

I would disagree with the statement that serial dilutions should not be used. If you use "good laboratory technique" and a Class A glass pipet, you can get excellent linearity. I have done this many times. But the other replies are also correct, if your technique is bad or your pipets are not calibrated, then you will have calibration problems. A combination of serial and parallel dilutions is often the best approach.
Merlin K. L. Bicking, Ph.D.
ACCTA, Inc.
5 posts Page 1 of 1

Who is online

In total there are 21 users online :: 0 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry