Friday: Changed Helium gas cylinder as was running low.
Put long run on over weekend, During run %RSD deteriorated and became erratic. Performed Tune before and after run;
AutoTune prior to run AutoTune after run
Repeller 30.12 30.63
Peak widths 0.60/0.61 0.60/0.61
69 ABS: 467573 388924
EM Volt: 1576 1576
Water: 6.77% 3.95%
N2: 2.67% 24.33%
O2: 0.46% 0.18%
CO2: 0.39% 0.08%
This would appear like a contaminated Helium Cylinder. Changed to a different cylinder and no improvement. Changed back to original low cylinder and still high N2, low O2.
Can anyone explain what’s happening? It’s not a typical leak as the N2:O2 ratio is not 4:1. It now seems it’s not a contaminated Helium cylinder. What other possibilities are there?
Any advise, ideas or suggestions would be much appreciated.