Dear all
i work with API 3000 and from 6 month till now a strange phonomena appears: contamination or intefrence peak appear at the same tR of my analyte even if blank sample from plasma or even solvent only the peak still appears.
when i moved to another analyte there still contamination with the new analyte.
what is strange is when i increse LLOQ the contamination peak increases

If I understand exactly what you are saying it sounds like a curry over effect that might be coming from your injector, valves etc...

I agree with Kostas to see the same effect at the same time in two separate analyses would imply that the problem lies in the LC-MS system and not with the chromatography.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry