I am facing a serious problem with the Waters Empower II software recently installed supporting two new Allaince+2998 HPLCs. I have communicated the problems to waters. There unability to solve this issue till date (I am following up for more than a month now) gives a strong hint that they are as flummoxed as I am.

the mail that I sent to the waters people is appended below.
"We would like to highlight some of the problems we are facing with the new PDA 2998/ empower software.
1. Theoritical Plates
This is very significant. The same sample analysed using the same mobile phase/ column on the same day is showing USP plates half or less in the new systems when compared to the older 2996 or 2487 systems. Both the software are empower 2.
2. Peak Purity
The peak which is showing perfect homoginity in the 2996 is being shown as impure in the 2998 detector. One of the causes may be the lower noise in 2998. However we need to understand if there is any issue when a method from 2996 is transferred to 2998.
3. Apply method set vs. Integration giving different results
This is very intriguing. Let’s say we have chosen a bunch of data from the “injectionâ€