ebruu » Mon Jan 28, 2008 9:15 am
Hi,
Blank solution preparation: To 50 µl acetonitrile, 50 µl 0,2 M borate buffer solution, pH : 8, was added, followed by addition of 50µl 500 µg/ml FMOC-CI solution resulting mixture is vortex for 10 s. The derivatization was performed at 50 C for 20 min in hot air oven.
Chromatographic seperation : I have used for chromatographic separation Hypersil MOS C8 column (5µm partical size, 200 mm x 4.6 mm i.d.)
Mobile phase is consist of methanol : water ( 20:80, v:v) and flow rate is set at 1 ml/dk
Prior to analyse, the column is flushed with the mobile phase for one hour and the baseline is observed stabile.
I have sent two blank chromatograms, both of two blank chromatograms are the same day.
May be not stable in hot air oven temperature?
Thanks