Any help would be really appreciated.
Edit: We have an agilent 5975 MS
Advertisement
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
I run the E2M2s on all my 5973/5975s and the volatiles instruments have a molecular sieve trap at the entrance to the pump, which keeps moisture down in the oil and also helps prevent any oil vapor from getting back up the line.I lucked out and there was nothing on the source/quads/detector. I had a 73 where the rough pump regurgitated into the diff pump and the diff pump into the chamber a bit. there was a residue on the chamber walls and some crud inside the triode guage tube. I got that all clean, replaced the foreline pump with an E2M2 with a trap so that won't happen again. Just methanol and MeCl2 worked for me. If it is just diff pump fluid it is glycol based and should be soluble in methanol.
Only makes it worse if you have an air leak. Otherwise it will help bake off any high boiling compounds that might have accumulated on the source. It also makes it get rid of water quicker after having the analyzer open.I'm worried I might be facing this problem as well, does the "MSD Bakeout" method do anything to help or will that just make things worse