IC : uneven baseline using Dionex 5000 RFIC system
Discussions about IC and related topics
I ran a seven anion standard mix and the baseline of my chromatogram is not all even. For each of the 7 individual peaks in the standard, each peak would start to rise at 0 uS but as each peak goes back down, the peaks will drop a bit lower than 0 uS and then would slightly go back up again to 0 right before the next peak would come out. Therefore, between each peak there would be an uneven baseline. Also, when I ran this standard about a month ago, I was getting a stronger signal for each of the anion peaks. Would anyone have a suggestion on what I could try to fix this or how I could narrow down what the issue might be? I have a Dionex 5000+ RFIC system, that has a suppressor, a continuously regenerated anion trap, and an added carbonate reducer. I am also currently doing auto-injections.
Thanks for any suggestions.
Hi Kuju,
Welcome to the forum. Based on what you have described- negative dips immediately after all analyte peaks and overall reduced response- it sounds like your suppressor capacity may be slightly depleted. Please try the chemical regeneration described in section 5.2.1 (steps A-C on page 77) of the suppressor manual (link here).
Best regards,
John Guajardo
Product Manager
Thermo Fisher Scientific
In total there are 23 users online :: 0 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)