- Posts: 2
- Joined: Wed Mar 20, 2019 5:59 pm
Advertisement
Ion Pairing reagent/Phloxine B
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
3 posts
Page 1 of 1
I'm testing for Phloxine B using 55% ACN with 0.03M TBAB at 548 nm. I'm getting peak splitting, could this be caused by too much ion pairing reagent, and what are the effects of using too much?
-
- Posts: 775
- Joined: Mon Dec 01, 2008 9:59 pm
How much you are injecting and what is the retention time? You might be seeing effect of large volume injection. Also, mixed-mode columns will allow you to avoid the use of IP reagent (it is attached to the surface of silica)
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
-
- Posts: 2
- Joined: Wed Mar 20, 2019 5:59 pm
I'm injecting 20 microliters. I noticed that if I use 65% with 0.01M TBAB, the peaks look completely different, and I cannot tell if it's the ACN or the TBAB.How much you are injecting and what is the retention time? You might be seeing effect of large volume injection. Also, mixed-mode columns will allow you to avoid the use of IP reagent (it is attached to the surface of silica)
3 posts
Page 1 of 1
Who is online
In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 16 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 16 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science