Dear reader
I have a vitamin A+D drop sample and I want to assay vitamin A$D . I use C18 column as stationary phase and mixture of CH3OH/ACN/ Hexane (46.5/46.5/7) as mobile phase and THF/ ACN (1/1 ) as diluents. AUC for vitamin A std (about 2200) but in the last 2times it is reduced to about 70. I checked the concentration , wavelength, mobile phase, diluents, but I can’t understand what happened.
Please Don’t hesitate me

Try injecting a sample or standard that contains more than one peak. If the response for all the peaks is lower than expected (by the same proportion), you probably have a malfunctioning autosampler.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Hi Sara

Do you have to separate D2 from D3?
3 posts Page 1 of 1

Who is online

In total there are 34 users online :: 3 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Google [Bot] and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry