Please Help me,
My column was left without solvent for more than one month. It is dry. It there a solution to restore the column package?
Thanks very much.
Rachid Ganga

All you can do is try to pump some mobile phase into it and hope for the best. Try to use a MP that is at least close to whatever was last used.

It may not be as bad as you think - it takes quite a while for an uncapped column to dry out.
Thanks,
DR
Image

Plus, don't give up to easily.
Follow DR's suggestion and start with the solvent that has been used last. There are some circumstances, where this may not yet recover retention. If this is the case, flush the column with 5 column volumes of water followed by 5 column volumes of acetonitrile. Then go back to the intended mobile phase.
3 posts Page 1 of 1

Who is online

In total there are 34 users online :: 3 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Google [Bot] and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry