I'm having a bit of a problem with a separation I've developed. I have developed a gradient, using the scouting method from Tom Jupille, to separate the following compounds:
- maleic acid
- phthalic acid
- o-toluic acid
- benzoic acid
- citraconic anhydride
- phthalide
A = 15mM KH2PO4, adj to pH 2.45 w/ phosphoric acid
B = ACN
flow rate = 2.0 mL/min
Time %A %B
0 100 0
1 100 0
9 65 35
12 65 35
Post time: 5 min
Column: Waters XBridge 4x150mm C18 3.5u @ 30°C
Detection: UV @ 210nm
and here is a 25uL injection of water on an Agilent 1100:

I replaced the eluents, containers and inlet frits for both eluents. Since I was using a previously used guard column, I removed it to see if that was the problem. Same results.
I've read here and there in this forum about problems with acidic phosphate buffers and acetonitrile. Should I try a gradient with methanol instead?
PS - Pardon my poor gradient table formatting. I'm not sure how to make a nice orderly table.