Most of the resources that I have found only discuss introducing liquid samples to the GC. How can I connect my Mixed Gas Instrument to the GC and control injection of samples?
Thank you!
Advertisement
Discussions about GC and other "gas phase" separation techniques.
I am new to the GC world, so I may be have just misunderstood. How can I connect my mixed gas instrument to the switching valves? What is the physical connection that I will have to implement?Probably the easiest way would be to use a switching valve and a 1-2ml sample loop. Let the effluent of the permeate flow through the valve and loop then switch the loop into the carrier flow and allow the carrier to sweep a few loop volumes onto the column, then reset for next sample.