We acquired Dionex DX-120 and we cannot find part number of the Rheodyne injection valve, originally mounted by factory.
Could you tell us what Rheodyne valve do you use in your DX?
What is its part number?
Thank you a lot,
Mykeil
This system is pretty old so you will be lucky to find a company selling new rheodyne valves. however Thermo sell rebuild kits.
https://www.thermofisher.com/order/cata ... uct/061947
or it might be worth contacting them to find out if they have the part.
I once worked in a lab where they used the injector, pump, column oven and autosampler from a Agilent 1100 and passed it through an Dionex DX-120 detector. we had a universal chromatography interface (UCI 100) that allowed the chromeleon software to pick up the signal form the DX-120.