Gafshar » Tue Nov 27, 2007 5:08 pm
Here goes, this is something I should have done from the beginning,
so I am using this fiber :
http://www.sigmaaldrich.com/catalog/sea ... ELCO/57301
for SPME, with FID, to detect benzene, with concentration ranges of 0.5-50 mg/L, on a CP-3800 GC and a CP- 8400 autosampler, with the following as the program set-up:
Autosampler:
Adsrob time: 5 min
Desorb time: 1 min
Sample Depth: 10%
Solvent depth: 0%
Oven:
45 degree, total hold 0.50 min
100 degree, total hold 7.38 min
200 degree, total hold 10.38 min
So this is what i was using for a good two months, i would have to switch columns and spme liners, and switch from standard to spme mode, and eventually calibrate the autosampler's position both on the vial and the injector port, and one day it started not working. these were the vials and caps i was originally using:
Agilent vial:
http://ecat.fishersci.ca/(cpvxns553coy3 ... id=3011009
Agilent Cap:
http://ecat.fishersci.ca/(hd1xewvtt4d13 ... id=3278399
So by process of elimination, it was decided that it must be the caps, for i had two set of caps with with different color septa, so that was eliminated, we thought it might be a specific position, but that was not the case either, and as suggested here, I switched the whole vial and caps to the following:
supelco:
http://www.sigmaaldrich.com/catalog/sea ... ELCO/27532
However, that is still a problem, in fact it takes longer for the fiber to break when using the agilent vials and caps, it runs about 30-40 samples before it goes astray. What i have noticed is that either way, the whole thing starts bending after it sits on the caps and wants to pierce, so it seems as if the pressure that is exerted on it to pierce is too much or something. so eventually as it keeps on doing that with all the caps, the fiber breaks from the inside and eventually its metal holder bends and breaks. So I recalibrated such that the sampler would sit about 0.5 cm above the caps, to see what happens, and what happened instead was that when i started to run the program, the autosampler would still actually sit on the caps, rather than staying above it about 0.5 cm as calibrated. So right now my final conclusion is that the gc itself needs to be aligned and serviced.
hope this cleans things up a bit. thanks again for your help and suggestions, keep them coming
Ghazal