PHOBIUS » Tue Nov 27, 2007 11:07 am
Biological samples (mainly extracts of leukemia cells) were only a small part of those 200 injections (max10%) and the extraction buffer was the same as the initial mobile phase of the gradient so I think, the samples were not the problem here.
Mainly (90%), I was injecting standards (trying to validate a method by measuring intrabatch and interbatch reproducibiliy, LOD etc.). The pH of the buffer was 7,00. The pH stability of the column (Synergi Hydro C18) is 1,5 - 7,0. The local dealer told me, that using pH about 7 is absolutely safe. well, the reality prove him wrong.
The next time I use rather a Synergi Fusion column, which also can be run under 100% aqueous conditions and has a pH stability 1.5 to 10.
I hope, this Synergi fusion sustain more than the Hydro one.
Thanks for your advices.
To Bryan Evens: What is the pH stability of the Unison UK-C18 (3um) and have you tested how much injections can you make without loss of retention or peak broadening using exactly the same conditions as here:
http://www.imtakt.com/TecInfo/TI133E.pdf
And are these columns more or less expensive than Synergi Hydro/Fusion?