In general what would contribute more to the "selectivity" of a fairly complex typical pharmaceutical RP-HPLC type impurity analysis - would it be contributions from the column or mobile phase contributions combined with gradient settings? My gut feeling says the selectivity contribution from the mobile phase and gradient settings far outweighs the contribution from the column alone. If this is the case, is there any real value in multiple column screening for early phase method development work? Would it make more sense to just stick with a common and garden C18 column and look at which organic mobile phase gives the greatest indication of peak selectivity? Then you would just develop and model your gradient separation using DryLab or other chromatography modelling software.
Also as an aside, is it possible to measure the contribution to selectivty of the mobile phase and column separately? I suppose this type of investigation would need to be run under fixed conditions which removes the possibility of enhancing selectivity by altering the gradient conditions.