usp calculation

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hi, I am doing a test for the calculation of Lidocaine according to the USP. Part of the calculation states Ru/Rs where Ru and Rs is the ratio of the peak area of the standard and sample. There is internal standard in both the standard and sample. Should I take the ratio of the active to internal standard in both the standard and sample and then take the ratio of the areas, or is it just the ratio of the active peaks it means in both the standard and sample.
Thanks Liv

You should use the ratio of the area of the active substance to the internal standard to do the calculations.
Otherwise the internal standard is of no use.

Bart

Here is a useful link explaining the use of internal standards.

http://www.lcresources.com/wiki/index.p ... alStandard
No Tswett
3 posts Page 1 of 1

Who is online

In total there are 23 users online :: 2 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Majestic-12 [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry