I used ESI negative mode and MRM scan. Plasma sample was extracted using MTBE. Retention times were stable and chromatography were good for the first and second injection.
Thanks.
Advertisement
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
Thanks for your reply. This two injections were not consecutive. this two injections are same calibration sample but injected twice before and after unknown samples. There are about 10 - 20 unknown sample between the first injection and the second injection. I did twice calibration sample injection in my method to monitor the stability of instrument. I found the signal intensity of the second injections were much higher than that of the first injection. I checked carry over and carry over was less than 0.1 %. thankswhat happens if your third injection is a blank?