When preparing an HPLC method (UV/Vis detector), one will usually determine the response factor through a calibration curve.

Can one use Beer's Law instead? A = Ecl and get El as the constant instead of RF?

Does this need to be done with every component, like in GC-FID or is it universal for the detector once determined?


Thanks!

Roxanne.

Beer's Law gives you the concentration that is in your flow cell when you take the measurement. Unless the entire peak is in the flow cell at one time, that doesn't directly tell you how much analyte came out of the column (that's why you integrate to get the area under the peak.

Different compounds have different UV spectra, so the calibration plot will be unique for a given compound. Different instrument/column combinations have different amounts of band broadening, so the calibration plot will be unique to a given system. And slight differences in mobile phase composition can make peak areas (and UV spectra) shift, so the calibration plot will be unique to a given day.

Bottom line: you have to calibrate the response factor for each compound on each system each day.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 24 users online :: 2 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Google [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry