Advertisement

Detection of PCl3

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hello,

Is it possible to detect PCl3 in GC with a NPD detector? I don't know if a NPD also needs a carbon it the analyte.

Thanks for Your information

Flower

Emsman,

Have not done it personally but you should be able to from what I understand about the detector. Phosporous is thought to rely on PO/PO2 formation which does not require hydrocarbon (Detectors for Capillary Chromatography, ed. H. Hill.) The down side, NPD's react very poorly to the presence of chlorinated solvents like DCM so this may work against you. If in doubt, ask the man who probably knows the most about NPD, Paul Patterson at detplp@aol.com.

Also should be do-able by PID or argon ionization detector. IP = 9.9 eV so should be able to do sub-ppm at a minimum although not as selectively.

Best regards.

I wonder if ECD would also work? I have never tried this, just a thought.

GCguy
GCguy
3 posts Page 1 of 1

Who is online

In total there are 113 users online :: 1 registered, 0 hidden and 112 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Ahrefs [Bot] and 112 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry