Let me know the following from the following column specifications: Capillary column AB 624, 30M, 0.53mmID,3 micron and
DB 624 30m, 0.32mmID, 1.8 micron

1) what are the advantages and disadvantages of Capillary column AB 624, and DB 624
2) Is there will be any similarity in both results
3) Which is the most preferred from the above for the solvents Cyclohexane and Methanol

swamyg444,


Both columns have the same stationary phase. The phase ratio for both columns is the same (calculated with Agilent's GC Method Translator);

You should expect therefore similarity between the results gained from both of them
The difference is that the megabore column AB 624 (i.d. 0.53mm) have greater loadability than DB 624 (i.d. 0.32mm). In other words, if you want to gain similar peak heights/areas with both columns, you should inject greater volume of the standard/sample solutions on the megabore (compared to the 0.32mm i.d. column)

Best wishes
2 posts Page 1 of 1

Who is online

In total there are 131 users online :: 1 registered, 0 hidden and 130 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Ahrefs [Bot] and 130 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry