I have to make injections in porapakQ-TCD with a sample with CO2, CH4, H2, and it maybe contains small quantities of H2S. Is H2S poisonous for porapakQ column or TCD? which temperature must I set the column for H2S to elute?
-
- Posts: 584
- Joined: Mon Sep 13, 2004 8:22 pm
tolisfoto,
H2S will be OK on a PorapakQ column and a TCD. An old Chrompak application shows on a 9 Ft PorapakQ column at 50'C, and helium carrier gas at 30 ml/min, CO2 has a RT of 1.5 mins and H2S has 5.2 mins.
Gasman
H2S will be OK on a PorapakQ column and a TCD. An old Chrompak application shows on a 9 Ft PorapakQ column at 50'C, and helium carrier gas at 30 ml/min, CO2 has a RT of 1.5 mins and H2S has 5.2 mins.
Gasman
2 posts
Page 1 of 1
Who is online
In total there are 126 users online :: 1 registered, 0 hidden and 125 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Ahrefs [Bot] and 125 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Ahrefs [Bot] and 125 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science