Chromatografic parameters:
X-bridge RP18 shield column
PDA 210 nm
Mobile phase A: Water : Acetonitrile 90 : 10
Mobile phase B: Water : Acetonitrile 20 : 80
Linear Gradient: 0 min : A:90 B:10
30 min : A:10 B:90
Issue : peaks appear from 20 min to 30 min
The peaks remain without injecting nothing so they do not come from the blank
They remain after total cleaning of the HPLC with acetonitrile, cleaning of all the material with acetonitrile, use of HPLC grade water from Merck or use of water filtered through an other X-bridge RP18 column.
I try also to extend the equilibraton time and the peaks areas increased but it can come from water or acetonitrile since the initial conditions are 90 water and 10 acetonitrile.
The last suspect is acetonitrile but it is a HPLC grade from Merck. I try other batches and an other supplier: same problem, same peaks.
Desesperated, I change acetonitrile by methanol keeping the same conditions: no peaks. Does it mean that the water is ok ???
Someone can explain this to me because I do not know what to do...
Thank you