Effect of pH on Stationary Phase

Discussions about GC and other "gas phase" separation techniques.

7 posts Page 1 of 1
Hi to all
My Quearry is .
What is the Effect of Sample Solutions pH on the Polar and Non Polar GC Columns Stationary Phase with the temp.Programming.( Like 100% DimethylePolysiloxane , 5% Phenyle-95% DimethylePolysiloxane , 35%Phenyle-65% DimethylePolysiloxane , 14%Cyanopropyl Phenyle-84%DimethylePolysiloxane , 50% Phenyle-50% DimethylePolysiloxane) :?:

Anything that is basic pH > 7 will quickly destroy a silicone based phase.

Organic acids generally do not harm silicones. Mineral or oxidizing acids will do harm and are not recommended.

Try not to use water with any silicone phase of any type.

best wishes,

Rod

Thanks for this
Actually My Sample is A Solution in Ethyle Acetate and Having Basic pH Near to 8-9.
We Found A Hump During Analysis. is It Due to Basic pH .or Due to Stationary Phase Problem.
My Column is DB-17 ( 50% Phenyle+50% Dimethyle Polysiloxane)

question?

why is the ethyl acetate basic? What base does it contain?

When I know that I can answer your question.

There should be no hump in ethyl acetate.

And a possible answer to your question:

It may be both if the first is true.......... since you injected your sample on the phase it may be damaged.

best wishes,

Rod

Re

Thanks ROD
My Compound is a mix. of Sec. Amine + Valaraldehyde + Cyanohydrine Solution is Given in 20 % Ethyle Acetate or in 20% MeOH.

I suspect your 'hump' is a product of your interesting mix. Aldehydes almost always contain traces of the oxidation product (acid) which would be volatile but not chromatograph well ( a hump ). Aldehydes are also know to polymerize (dimer).

But it is most certainly possible that the secondary amine could have (could be reacting) reacted with the fused silica surface or silicone phase.

I suspect the former is more likely than the latter.

Like Bruce says........ Keep having fun.

Rod


Try not to use water with any silicone phase of any type.
I've always believed this (initially starting out in the environmental field). However, once entering the pharmaceutical field I've found that the USP loves injecting water (either by headspace vapor or direct inject) into 624 phase (G43) columns. Good ol' USP <467> is mainly written towards water soluable materials.

How long does it take to eventually destroy the phase in one of these columns with, let's say, 10-15 water headspace injections (1mL) a day?

Burt
7 posts Page 1 of 1

Who is online

In total there are 101 users online :: 0 registered, 0 hidden and 101 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 101 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry