kf auto titrator calibration

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
can anybody tell me why we are using sodium tartrate di hydrate during KF auto titrator calibration?

Because it is not hygroscopic and contains a known, stable amount of water.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

As Tom said, it's a primary standard ( water content 15.66% mass/mass ), it's readily soluble in methanol, plus it's the suggested calibrant by the USP for samples with trace amounts of water ( < 1% ), although you can also use water.

Bruce Hamilton)

THANKS BOTH OF YOU.

We use water as the KF calibrant. We are usually looking at higher levels, from a few percent to 90%. We use a gas-tight syringe, 100 ul.