CN column for Benzalkonium chloride assay
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
7 posts
Page 1 of 1
Hi there. Currently I am assaying Benzalkonium chloride using a radial compression Nova-pak CN type cartridge column, and would like to change over to using a stainless steel Waters CN column. Cetylpyridinium chloride is used as an internal standard. I tried to use the current mobile phase on the stainless steel column, which is 1000mL MeOH, 20 mL glacial acetic acid, and 1.0g hexanesulfonic acid sodium salt. It gave me horrible peak shape, using 10uL injection volume, and 1mL/min flow rate. In an attempt to rectify this, I added 200mL water to the phase. This significantly improved the peak shape, however it also caused the peaks to elute too close together. Any advice to help me get better resolution? Thanks, James
-
- Posts: 583
- Joined: Thu Feb 23, 2006 3:15 am
CN columns are not reproducible. They should be a last resort.
Benzalkonium chloride has been retained on Unison UK-C8:
http://www.silvertonesciences.com/files/TI197E.pdf
Benzalkonium chloride has been retained on Unison UK-C8:
http://www.silvertonesciences.com/files/TI197E.pdf
-
- Posts: 583
- Joined: Thu Feb 23, 2006 3:15 am
Correction - a more appropriate description would be that CN columns
are "unstable" as opposed to "not reproducible".
are "unstable" as opposed to "not reproducible".
-
- Posts: 179
- Joined: Tue Nov 09, 2004 5:04 pm
JFrancis,
I don't remember all the details of the benzalkonium chloride analysis method that I used a while back at a previous company, but I will give what details I can remember.
I think that we used a Zorbax SB-CN column. The mobile phase may have been a simple methanol/water mixture. I can't recall if we used a buffer or an ion pair reagent, but I don't think that we did.
We did not use an internal standard. We did get the reference standard from the USP.
You may already know, but the benzalkonium chloride does not come as a pure material. The material contains a mixture of compounds with different alkyl chain lengths from C12 to C20. If I remember correctly, the USP reference standard contained compounds with the C12, C14, C16 and C18 alkyl chains (maybe C20 also) while our material had only the C14, C16 and C18. You need to quantitate using peak area grouping for all of the benzalkonium peaks, in the standard and sample.
So your chromatogram should give you 3 or 4 peaks for the benzalkonium chloride. Perhaps, you had bad peak shape due to incomplete resolution.
Sorry, I can't be more complete on the method details. It has been a while and I can't just go look it up. As the USP has a reference standard, there may be a monograph method that you can look up. I don't have a USP to search.
Regards,
Dan
I don't remember all the details of the benzalkonium chloride analysis method that I used a while back at a previous company, but I will give what details I can remember.
I think that we used a Zorbax SB-CN column. The mobile phase may have been a simple methanol/water mixture. I can't recall if we used a buffer or an ion pair reagent, but I don't think that we did.
We did not use an internal standard. We did get the reference standard from the USP.
You may already know, but the benzalkonium chloride does not come as a pure material. The material contains a mixture of compounds with different alkyl chain lengths from C12 to C20. If I remember correctly, the USP reference standard contained compounds with the C12, C14, C16 and C18 alkyl chains (maybe C20 also) while our material had only the C14, C16 and C18. You need to quantitate using peak area grouping for all of the benzalkonium peaks, in the standard and sample.
So your chromatogram should give you 3 or 4 peaks for the benzalkonium chloride. Perhaps, you had bad peak shape due to incomplete resolution.
Sorry, I can't be more complete on the method details. It has been a while and I can't just go look it up. As the USP has a reference standard, there may be a monograph method that you can look up. I don't have a USP to search.
Regards,
Dan
-
- Posts: 974
- Joined: Tue Jan 11, 2005 11:37 pm
The Acclaim Surfactant column gives excellent peak shapes and resolution for quaternary-amine surfactants, including benzalkonium salts. No ion-pairing reagents needed. Look at page 5 of this data sheet: http://www1.dionex.com/en-us/webdocs/25 ... et_V24.pdf
Mark Tracy
Senior Chemist
Dionex Corp.
Senior Chemist
Dionex Corp.
-
- Posts: 3
- Joined: Mon Sep 24, 2007 1:19 am
Thanks a lot for your help guys. The existing method with the radial compression cartridge only separates one peak for all of the homologs of Benzalkonium chloride, and I would like to keep this as being the case, if at all possible, rather than split them all up to get resolution for all of the homolog peaks.
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
There could be an equilibration issue. The Radial-Pak Cartridge is dry, when it arrives in your hands. The steel column is shipped wet, if I recall correctly, it is shipped with a mixture of acetonitrile and a salt solution used to test the column. Before you will be able to use the column in 100% organic, you need to remove the salt. Most likely, you have done this now by flushing the column with an eluent containing water. I would go back to the original mobile phase and see if the separation is now what you had anticipated.
The CN columns hydrolyze in general, so I am not a fan of them, but they should give you reasonable results in the non-aqueous medium that you are using.
The CN columns hydrolyze in general, so I am not a fan of them, but they should give you reasonable results in the non-aqueous medium that you are using.
7 posts
Page 1 of 1
Who is online
In total there are 8 users online :: 1 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 7 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 7 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science