If you use a regression line through the points used in the excel file of JA, combining the 2 injections, plotted versus the concentration, without any weighing, you have a non significant intercept (t-test).

But in an ideal world, you first have to prove that the data is homoscedastic.
If your data is heteroscedastic, you have to use weighing or log, x^2, 1/x.... to make sure your data is homoscedastic.

If your intercept is significant, you can use a calibration curve instead of a single point calibration. If you use a log-log scale, your residuals reduce by a great amount.

If y=ax^b and you use log on both sides of the equation, then you have

logy = loga + b logx

so the intercept in logy ~ logx curve (loga) is the log of the slope in y ~ x curve. The slope in log y ~ logx curve (b) should be close to 1.00(between 0.98 to 1.02) if the ralationship between y and x is linear. Otherwise y ~ x curve is considered quadratic. In your case, the slope was 0.9925.

If y = ax^b + c, which means y ~ x curve has a significant intercept,
then logy = log (ax^b + c), there is no linear ralationship between logy and logx.
17 posts Page 2 of 2

Who is online

In total there are 8 users online :: 1 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry