gas phase internal std

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Dear Chromatographers.

We are implementing the EPA TO-17 method. It calls for gas phase internal standard transfer onto each sorbent tube.
One of options is to transfer from cylinder to a loop and then automatically to the sorbent tubes in the TD instrument. Nevertheless for the moment We can´t implement this setup. The question is, Could we use instead a pressure regulator, a syringe adapter and a gastight syringe to take an aliquot of the gas and inject it manually onto each tube?

Best regards.
I don't use this method but I don't see how what you're doing with the syringe is any different than what is done with the loop. It's all about getting a known and reproducible amount of the standard on the sample tube. It seems to me that you can accomplish that either way. However, I don't know what an auditor might say......
That´s true, The auditor´s point of view about this.
He would ask for a statistic demostration of equivalence.

Thank you rb6banjo.
3 posts Page 1 of 1

Who is online

In total there are 26 users online :: 2 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: GRAMIN88, Semrush [Bot] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry