FMOC-Cl cleanup for LC-MS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
I'm doing a project with FMOC-Cl derivatization of glyphosate. We started off using a 500ul of 2mg/ml FMOC-Cl concentration which is added to 500ul of extracted glyphosate and borate buffer. We would filter then inject the sample on a 2.1x150mm XBridge column with 5mM Ammonium Acetate in water Water and Acetonitrile as the mobile phase.

In reading the literature, I see a wide variation in the recommended FMOC-Cl concentrations for derivatizing. Thus, I was compelled to investigate how the results would change if I increased the FMOC concentration. The highest concentration that I tried is giving satisfactory results. I'm now using 500ul of 25mg/ul FMOC-Cl. I'm still injecting the derivatized samples without further cleanup after derivatization.

I've seen where a researcher working with peptides published that they do a 80%pentane-20%ethylacetate partition to remove the FMOC by products. Surely this is also applicable to FMOC-glyphosate, and I've tried this but my peaks look horrible after the partitioning.

I added 2mls of the pentane:ethylacetate to my 1ml of sample, borate buffer, and FMOC-Cl mixture, then vortex for 5 seconds. Let sit for 10 miunutes then use a pasture pipette to remove the lower aqueous portion and put in a vial, then cap and inject. Do I need to centrifuge step after doing the vortex?

Or, since the chromatography still looks ok, should I just not worry about the FMOC by products on the column. I've been increasing the acetonitrile in the mobile phase to 100% for 10 minutes to flush the column after each injection, and so far so good.

Any recommendations?
Try to use a polymer supported primary amine to scavenge Fmoc Cl. Then you would just filter to remove the beads and inject your sample as previously.
Try to use a polymer supported primary amine to scavenge Fmoc Cl. Then you would just filter to remove the beads and inject your sample as previously.
Thanks Ericf, I'll give that a try.

A piece of equipment is down so I had time to do this comparison. I went ahead and did a comparison between injecting some low range standards 50-200 ppb after a single pentane:ethylacetate partition. Compared with no clean-up of the derivatized sample.

My straight injection results are probably acceptable, however, after the pentane:EA clean-up, I had nice peak shape and significantly improved signal to noise ratios; s/n's increases were 10-30%.

I like the sound of the amine terminated beads. That might be easier.
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry