Dear all
while during the method development for LC MS MS, which is correct the process
1. chossing column first
2. chossing mobile first
please explain me simply sir
regards
venkateswarlu p
Scientist,
Vimta Labs.
Cherlapally Hyderabad.
India
How ru? while devloping the lcms method
chosing the mobile phase is first then you can go to the column.

Cause normally all the compounds will elute in C 18 Column especially pharmaceutical products.

Depending on the hydrophobicity, ionizability (basic, acidic...) etc of your compound(s) you first choose the column that will tentatively retain/separate your compound(s) and then a MS compatible mobile phase (volatile additives, if salts are added (always volatile) it would be better to use them at less than 20 mM) that will provide the retention/separation you desire.
3 posts Page 1 of 1

Who is online

In total there are 45 users online :: 0 registered, 0 hidden and 45 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 45 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry